Could someone tell me if it is possible in fluke 1663/1664 to activate the option to carry out an insulation test without having the finger fixed on the test button? I am not referring to the differential test, which I know can be done by detecting voltage and not pressing the "start". I only mean insulation test... or maybe it's not possible...? (it is not in the manual)
The only conclusion I can draw is for the safety of the operator... Other manufacturers, which are many, do the test by pressing the insulation test once...
 
I have testers that take both approaches. Metrel is press and let go and it takes samples over a couple of seconds. Meggers are press and hold.

The advantage of those that keep the current running while you hold is that you can see if the reading is slowly improving or vice versa. Sometimes you can see lower IR which improves as you are physically drying out the cable by running current through it.
If anything, I consider the method your tester uses to be the superior method!
 
Last edited:
I have testers that take both approaches. Metrel is press and let go and it takes samples over a couple of seconds. Meggers are press and hold.

The advantage of those that keep the current running while you hold is that you can see if the reading is slowly improving or vice versa. Sometimes you can see lower IR which improves as you are physically drying out the cable by running current through it. I
If anything, I consider the method your tester uses to be the superior method!
Cleared up, thanks a lot Timhoward.
 

Reply to Insulation fluke 1663/1664 in the Electrical Tools and Products area at ElectriciansForums.net

Similar Threads

Hi I've inherited a very tight space, a KMF switch and some split-con for a sub-main to try and tidy up. It's a joyous non-restrained single...
Replies
3
Views
239
Im making this cct for fun. Ive seen it in a video on youtube, with different IC's. So I decided to adapt the cct to what IC's I already have. And...
Replies
30
Views
4K
I own a top floor tenement flat that I used to live in and then rented out after I married. It is currently empty whilst some work is being done...
Replies
0
Views
225
Slightly strange one this, replaced a few socket fronts in a dining room today, all wiring original but new plasterboard on a studded wall. Its a...
Replies
4
Views
1K
Hi all, New to the forum. I have been asked to look at this for one of our guys who's had an issue onsite after some electrical works had been...
Replies
4
Views
873

Electricians Tools | Electrical Tools and Products

Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.

We also have a Tiling Tools advice from the worlds largest Tiling community. And then the Plumbers Forums with Plumbers Tools Advice.

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top